E267 ERR_NR_GEN_ERROR
General NR failure source unknown top level reporting.
1. Cycle power to the unit. Allow the unit to go through its self-check. If the error clears,
the unit is ready for use.
2. If the error does not clear, perform a level-6 calibration of the system as described in
Calibration Levels on page 6-6.
3. If recalibration does not work, open the cover to the unit and inspect the controller
board. Verify that the board is securely seated and fastened as expected. Verify that
the ethernet cable is connected to the system.
4. If the unit still displays the error and/or the recalibration does not correct the issue,
replace the controller board and perform a level-6 calibration.
5. If the new board does not correct the issue, contact your local Covidien Technical
Service for further instructions (see page 9-4).
E288 ERR_NR_SENSOR_CLIP
DSP sensor clipping error.
1. Open the cover to the unit and inspect the controller board. Verify that the board is
securely seated and fastened as expected. Verify that the ethernet cable is connected
to the system.
2. Cycle power to the unit. Allow the unit to go through its self-check.
3. Perform a level-6 calibration of the system as described in Calibration Levels on page
6-6.
4. If the unit still displays the error and/or the recalibration does not correct the issue,
replace the HVDC board and perform a level-4 calibration.
5. If replacing the HVDC board does not correct the error, replace the controller board
and perform a level-6 calibration.
6. If the new components do not correct the issue, contact your local Covidien Technical
Service for further instructions (see page 9-4).
E289 ERR_NR_SENSOR_COMPARE
DSP sensor comparison error.
1. Using the data information found under Service -> Statistics -> Error List on the
display panel of the unit, use table below to help isolate the error.
2. Open the cover to the unit and inspect the controller board. Verify that the board is
securely seated and fastened as expected. Verify that the ethernet cable is connected
to the system.
3. Cycle power to the unit. Allow the unit to go through its self-check.
4. Perform a level-6 calibration of the system as described in Calibration Levels on page
6-6.
5. If the unit still displays the error and/or the recalibration does not
Matt
Report Post
You will also find the manual under documents scroll down you'll see it. click and down load.
Matt
Report Post
Hi,
Tihs is from service manual:
E267 ERR_NR_GEN_ERROR
General NR failure source unknown top level reporting.
1. Cycle power to the unit. Allow the unit to go through its self-check. If the error clears, the unit is ready for use.
2. If the error does not clear, perform a level-6 calibration of the system as described in Calibration Levels
3. If recalibration does not work, open the cover to the unit and inspect the controller board. Verify that the board is securely seated and fastened as expected. Verify that the ethernet cable is connected to the system.
4. If the unit still displays the error and/or the recalibration does not correct the issue, replace the controller board and perform a level-6 calibration.
5. If the new board does not correct the issue, contact your local Covidien Technical Service for further instructions
BR
KV
Report Post
Yep, the manual for the ForceTriad is located here
And the Service Manual is here
https://www.medwrench.com/documents/view/4748/valleylab-forcetriad-valley-labservice-manual
These E288 and E267 errors are known and fixable. So you are in good shape.