XGI-8 Gas anesthesia system - how to exchange the VAPORIZER
RE: Anesthesia Gas Machine Report Post
I wish I could help you. I can't even get the vaporizer off my unit to get it calibrated. And getting help from Caliper is IMPOSSIBLE.
Report Post
You might wish to contact VetEquip in California. Removal is not difficult but you cannot exchange this with a standard vaporizer. This particular vaporizer requires a special calibration under pressure, there are not a lot of vaporizer companies that are equipped to handle this. You also cannot put a standard vaporizer in its place so if a loaner is required you would need to acquire one from the service provider and not just one off of the shelf. There is usually no indication on the vaporizer that it has been calibrated for use in a unit that delivers a constant pressure to the vaporizer. This is usually the case if you find a flowmeter positioned downsteam of the vaporizer.
Rick Trexler, RVT, CBET