Sun Jul 21 2019

Error 13

Hello! I am Rodrigues, I have a polymobil plus X-ray has error 13, we checked, and we found the fuse 80A burned. we changed the fuse, but even so it appears err

   RE: Mobile Radiography Report Post
Tue Jul 23 2019Reply from MedWrench Mandy
MedWrench Mandy

Here are the documents for this product. Hopefully the Troubleshooting Guide will help.


Mandy Hernandez | Digital Marketing Coordinator
email me: mandy@medwrench.com
Connect with MedWrench:
Facebook
Twitter
LinkedIN

Report Post
Tue Jul 23 2019Reply from Tomelgy
avatar placemark

What version Polymobil? Plus, 2, 3 etc!

The 80A fuse is in the kV inverter, look around for another fuse from DC supply like 20A/40A depending upon model.

What voltage do you measure on 80A fuse during exposure? Should be around 450V DC

E13 is no kV.

Tom


Report Post
Sat Oct 05 2019Reply from Slymn
avatar placemark

Did you checked igbt modules. What number of skm400 or 300.   And tested with oscilloscope 