Hello! I am Rodrigues, I have a polymobil plus X-ray has error 13, we checked, and we found the fuse 80A burned. we changed the fuse, but even so it appears err
Sun Jul 21 2019
Error 13
Tue Jul 23 2019Reply from MedWrench Mandy
Here are the documents for this product. Hopefully the Troubleshooting Guide will help.
Mandy Hernandez | Digital Marketing Coordinator
email me: mandy@medwrench.com
Connect with MedWrench:
Tue Jul 23 2019Reply from Tomelgy
What version Polymobil? Plus, 2, 3 etc!
The 80A fuse is in the kV inverter, look around for another fuse from DC supply like 20A/40A depending upon model.
What voltage do you measure on 80A fuse during exposure? Should be around 450V DC
E13 is no kV.
Tom
Sat Oct 05 2019Reply from Slymn
Did you checked igbt modules. What number of skm400 or 300. And tested with oscilloscope