Thu Aug 08 2019
error 061
Hi every body, I have error 061 on bv pulsera c arm can some body help me to fixit ?
best regards
Thu Aug 08 2019Reply from Arthur
Hi,
I believe M061 is when you have a mismatched C with the stand. Do you have more than one system? Could you have the wrong monitor cart?
Art
Sat Aug 10 2019Reply from Gerardolobo
there is no service manual and there in not a second unit to do fast testings.
The cable was checked but it is ok.
If you have service manual send it please to g_fuentesg@hotmail.com
Thank you very much for your help.
...G.Fuentes
Mon Aug 12 2019Reply from lao
Sounds like the viewstation and the c-arm serial number do not match anymore. If you dont have another unit which was switched out by mistake it probally lost the information. You will need someone with philips BV scope login information to sync them back up.
Fri Dec 13 2019Reply from PMS