What should be the resistance range for a healthy heater element? Unit continues to blow overtemp control.
Thu Aug 08 2019
Heater element
Fri Aug 09 2019Reply from MedWrench Mandy
What is the manufacturer and model name?
email me: mandy@medwrench.com
Connect with MedWrench:
Fri Aug 09 2019Reply from Chiro_Bob
Mon Aug 12 2019Reply from Chris S
Somewhere between 20 and 100 ohms. Obviously should not be shorted or open.
Mon Aug 12 2019Reply from MedWrench Mandy
Here are the documents for this product that may be helpful.
Mon Aug 12 2019Reply from Statbiomed
The answer will be approx 14 ohms The reason the overtemp pops is because the water temp has exceeded 180 degrees or the overtemp sensor has gone bad
Tue May 05 2020Reply from Hot Papa
Would it be possible to have access to a wiring diagram ? I am having the same issue with overtemp tripping. Also where can I buy parts ?
Thank you
Tue May 05 2020Reply from Statbiomed