I am getting intermittent diluent sheath empty errors. Clearing the fault, refills fluid and background is good. Ran the low control, ok. The normal and high control wbc counts are high. I see blood in lines on the right side around 11,12,13 valves after a sample is run. Is that normal? Any input on high wbc on controls?
Thu Aug 15 2019
Controls and diluentsheath errror
Tue Aug 20 2019Reply from Paul Moyer
Try reseating the connections on the Diluent and Sheath chambers, also check too make sure all tubings are in their respective pinch valve, valve 11 and 12 could be part of the waste function, valve 13 I think is for HGB, I do not have a diagram so I am guessing as to which valve goes where.