mechanical ventilation

There are two devices dräger evita 2 dura. One is not working, the second is working. On one, the flow sensor is not calibrated, writes: calibration filed - measurement inop. On another device, this flow sensor is calibrated OK. Something with the device. Help advice.

   RE: Ventilator Report Post
Mon Aug 19 2019Reply from MedWrench Mandy
MedWrench Mandy

Thank you for your inquiry concerning Draeger. All product or service questions for Draeger devices can be directed to Draeger FREE of charge by dialing 1-800-4-draeger.


Mandy Hernandez | Digital Marketing Coordinator
email me: mandy@medwrench.com
Connect with MedWrench:
Facebook
Twitter
LinkedIN

Report Post
Tue Jan 28 2020Reply from Eng. Hamad
avatar placemark

There might be a problem with the flow sensor. It has to be replaced. 


Report Post
Tue Jan 28 2020Reply from mgbmike
avatar placemark

A few things to try. Swap flow sensors, if problem follows, then it's the flow sensor. If failure remains, swap exhalation valve. Flow calibration happens during inspiration. If the exhalation valve is leaking during inspiration, you can get this error. Lastly would be pneumatic controller PCB


Report Post