Service Manual
MedWrench is available to help people with specific questions and troubleshooting problems. Could you please be more specific with the issue that you are having with your instrument, and the MedWrench team will do their best to assist you in solving the issue. This will also help other community members who face similar challenges. Available documents can be found here.
email me: mandy@medwrench.com
Connect with MedWrench:
If you would like to share with the MedWrench community, please email your document to mandy@medwrench.com and we will upload it to the product page on your behalf.
Por favor será que puede compartir el manual de servicio del Arco en C siemenen arcadis advantic generación 1 gracias
MedWrench está disponible para ayudar a las personas con preguntas específicas y solución de problemas. ¿Podría ser más específico con el problema que tiene con su equipo? Esto también ayudará a otros miembros de la comunidad que enfrentan desafíos similares.
Hello!
Sir,
Hope you doing well!
Need help to fix a issue Siemens Arcadis Varic C-arm Error Code:5004
Kindly share Service Manual, Error Codes
Thanks & Regards
Zohaib
engr.zohaib54@gmail.com
The errors are not listed in the servcie manual, they are in the software disks.
What system generation1 (VB13C) or generation 2 (VC10)?
What is the letters in front of error 5004? BIC, MSY Gen etc. etc.
If its MSY_5004 then you have a generator error, check power supplies & the UZ voltage. Should be between 200 & 320V DC.
When does the error appear during standby or with radiation release?
Set to manual mod (40kV) & see if the error appears then also.
Could be a D1 CPU PCB, F1 of D2 PCB.
Tom
Hi Tom,
I get error codes 2006 and 2405. Any ideas what those error codes indicade? Appreciate your help in advance.
Best Regards,
Yahya
If the error is MSY_2006 then there was an issue with conne tion between PC & Main system D1 PCB. Check X10 pins for damage, PC conenctions and D1 PCB connections.
If MSY_2405 is similar to 2006, Main system interuption.
Tom