Error Code F 32

Hello Please provide us some solutions for Shimadzu x ray system (Portable X-ray)
Model: mux-100- Error code F 32


   RE: Mobile Radiography Report Post
Thu Sep 12 2019Reply from MedWrench Mandy
MedWrench Mandy

Have you check the service manual?


Mandy Hernandez | Digital Marketing Coordinator
email me: mandy@medwrench.com
Connect with MedWrench:
Facebook
Twitter
LinkedIN

Report Post
Thu Sep 12 2019Reply from Arthur
Arthur

This error means the measured filament current is more than 50% lower than set value.

I believe there is a manual for this device on Medwrench. I would start by attempting to calibrate the Tube current.

Art