cleaning recommendations
The reocommendations are in Chapter 6 of the instruction manual https://cdn.medwrench.com/eMan/equipmentNo/4495/bard-site-rite-6-instructions-for-use.pdf
6.1 Cleaning Procedures
To clean the keyboard, scanner, probe and A/C adapter:
1. Turn off the system.
2. Dampen a nonabrasive cloth with either warm water or rubbing alcohol.
3. Gently wipe the dampened cloth over exterior surfaces requiring cleaning.
Note: When cleaning the system and components, it is important to remove all particles or other
matter from all surfaces and crevices.
6.2 Disinfection Procedures
The Site~Rite* 6 Ultrasound System probe may be liquid disinfected by soaking it in Cidex* Plus 28
day solution.
Follow the solution manufacturer's recommendations for soak time necessary to achieve the desired
germicide level of activity.
Warning: Do not allow liquid to enter the scanner, A/C adapter, base power unit, probe connector or
probe port. Damage to the equipment may occur.
Warning: Do not attempt to sterilize the Site~Rite* 6 Ultrasound System scanner or probe with
ethylene oxide or heat sterilization methods. Damage to the equipment may occur.
Caution: When disinfecting the probe with a liquid disinfectant, do not soak the
probe cable, cable bend relief, probe connector or probe buttons.
Doing so may damage the probe.
Caution: Hot water (in excess of 113°F or 45°C) may damage the probe.
Caution: Use only Bard Access Systems cleaning and disinfection procedures.
Failure to do so may damage the device.
For a list of disinfectants recommended for use on the Site~Rite* 6 Ultrasound System and probe, contact
Bard Access Systems for the "Site~Rite* Ultrasound System Compatible Disinfectants" document