After 18th arrow the user interface wrote (xray room open) . Any help

   RE: Mobile C-Arm Report Post
Fri Oct 04 2019Reply from Arthur
Arthur

do you have manuals? I think this error is due to the 24V Interlock circuit being open.

 

Art


Report Post
Fri Oct 04 2019Reply from C12Guy
C12Guy

display stops at 18th arrow Check interlock path for open circuit.
Check cables and connectors Check INTLK_COMPLT* line to X-Ray Controller PCB for low. If high, interlock path is open.

DS1 LED int on the HV Supply regualtor that indicates when the 24V interlock is closed.

Room Control Tests
This test verifies that the outputs on the ROOM INTERFACE jack work correctly. This connector is located
on the Workstation's External Interface Panel at the rear of the Workstation.
Power Injector Test
WARNING: Performing this test generates X-rays. Follow safety precautions to prevent
exposure to X-radiation.
1. Connect the Interconnect Cable to the Generator if it is not already connected.
2. Apply power to the system and allow the system to completely boot-up.
3. Access the Mode touchscreen and select the Vascular Imaging Mode.
4. Connect an ohmmeter to the pins 1 and 6 on the ROOM INTERFACE connector.
5. Verify that the meter reads infinity (relay contacts are open).
6. Press the DSA footswitch and verify that after the syringe is displayed on the LCD screen that the
ohm-meter reads approximately 0 ohms (relay contacts are closed).
21
External Input/Output (I/O)
X-ray On Light Test
To perform this test the medical facility must use this feature and have an X-Ray On light on the
premises.
WARNING: Performing this test generates X-rays. Follow safety precautions to prevent
exposure to X-radiation.
1. Verify the X-ray On light cable is connected to the ROOM INTERFACE connector on the Workstation’s
External Interface panel.
2. Connect the Workstation's Interconnect Cable to the generator if it is not already connected.
3. Power on the System.
4. Press any X-Ray switch or pedal and verify that the X-Ray On Light illuminates while X-Rays are
generated.
Room In Use Test
To perform this test the Medical Facility must use this feature and have a Room In Use light on the
premises.
WARNING: Performing this test generates X-rays. Follow safety precautions to prevent
exposure to X-radiation.
1. Verify the Room In Use Light cable is connected to the ROOM INTERFACE connector on the
Workstation’s External Interface panel.
2. Connect the Interconnect Cable to the generator if it is not already connected.
3. Power on the System.
4. Press any X-Ray switch or pedal and verify that the Room In Use Light illuminates while X-Rays are
generated.
Door Interlock Test
This feature can be verified without connecting the Workstation to a magnetic reed door contact.
WARNING: Performing this test generates X-rays. Follow safety precautions to prevent
exposure to X-radiation.
1. Connect the Interconnect Cable to the generator if it is not already connected.
2. Power on the System.
3. Jumper pins 4 and 5 on the ROOM INTERFACE connector. This jack is located on the External
Interface panel at the rear of the Workstation.
4. Press any X-Ray switch or pedal and verify that X-Rays are disabled.