When i turn the device on it display E274 host self error so i enter to service ==>diagnostic ==>self test
I run self test and the RF generator fails
So what can i do to solve this problem?
Tue Oct 15 2019
E274 forctraid
Tue Oct 15 2019Reply from Matt D.

Have you contacted Covidien/ Medtronic/Valleylabs Tech support 800-722-8772? They are very helpful.
Matt D.
Report Post
Tue Oct 15 2019Reply from Valjak

Hi, this is from Service Manual:
E274 ERR_NR_CRITICAL_DATA_ERROR
1. Inspect controller board. Verify that the board is securely seated and fastened. Check the ethernet cable
2. Verify that the ribbon cables and FTSW cable connections are OK
3. Perform complete calibration of the unit
4. If the unit still displays the error, replace the system scanners and calibrate them
5. If this not correct the error, replace the controller board and perform complete calibration
Good luck