Problem restoring user settings
RE: Defibrillator Report Post
Have you checked any of the manuals?
One of these companies may be able to help:
Mandy Hernandez | Digital Marketing Coordinator
email me: mandy@medwrench.com
Connect with MedWrench:
Report Post
This device will need re-dux kit up-grade. The kit comes with new system board, printer, and power supply.
joey
Edge Biomedical
Report Post
I've seen the coin cell installed incorrectly and the positive contacts were on the negative side causing this issue.
We can help with this issue at ReNew BioMedical. Please call 844-425-0987 and ask for Neil
Report Post
Also wanted to add that the error codes will always appear after replacing the coin cell.
you have to clear the codes, clear the memory and power on your unit twice typically before the service light goes away.
seems as if you've done this as you mentioned all the time.
Report Post
only Stryker has the software.
we can offer trade in credits and put you in a recertified device for less than a software upgrade.
844-425-0987