Monitor cart not booting up
I don't know the Ziehm 7000 but thinking logically.
Could be graphic card in PC or cable from PC to monitor. If both monitors do not display any image/text then more likely graphic card.
Dark image with radiation correct?
What is the kV doing? is it hovering around 40/45kV or up at 110kV. What happens if you push kV up manually?
If you put Cu (or lead) in beam does the kV run up higher?
If the kV is tracking up/ depending upon the amount of Cu/Lead then the problem could be the TV camera cable coax broken. This depends upon how/where the auto dose take data/signal from to calculate the amount of radiation/dose required.
Tom
Then most likely your image intensifier power supply has a problem. Normally its supplied with 20/24V. You need to trackdown this.
Look at the II power supply it could have an power LED.
You could have a broken wire in the connecting cable from the electronics box up to the camera/II power supply on the C-Bow.
Okay thanks. Gives me a start.
You seem to really know your C-Arms. If I decide not to sink anymore money into them since they are very old and upgrade to something newer...what are your thoughts between a Ziehm solo, 2014 Siemens Compact L, or something like the 2005 Ziehm Vision, or Ziehm Quantum. OEC9600 is just as old as Ziehm 7000 so not really a great option.
Have you had any experience with the new Genoray Oscar from Korea ?
My issue is finding any good service techs for older Ziehms in the Phoenix area. Only thing they want to work on is the Ziehm Solo.
I don't know the Genoray Oscar from Korea systems.
The Zeihm solo & vision are pretty good systems, with IQ & patient Dose.
The Siremobil Compact although a good 'general system' if you want a system with really good IQ then go for the GE OEC Brivo or latest OEC One system.
Depending upon your application needs perhaps the up market Elite, more power, subtraction etc.
Thanks for the info. I'll keep investigating on my end. We only have very basic needs for the C-arm.