"Pressure set not fitted" on a Alaris CC pump
RE: Infusion Pump Report Post
Pressure set not fitted means, a set with pressure sensing disc isn't attached, sets like G30402M or MFX2294.
Also signifies the device bought is on dedicated mode (service access code 301), hence it is asking to insert/use set with pressure sensing disc.
Having a sensing disc the pump will be on variable pressure monitoring of 1mmHg steps from the scale 1mmHg to 1000mmHg.
Access codes and use
251 User Configuration Menu
(Drug Library, Bolus setting, Pressure setting but not for Type of set use)
376 Service Access Menu
301 Fully Dedicated (CC only) only dedicated set with pressure sensing disc.
302 Semi-dedicated (CC only) the device can use dedicated or not
GLacsi
Report Post
Sorry, I didn't see your replies, thank you so much for your help, GLacsi and Dany1981, appreciated!