OEC 9600

What needs to be done when the C reads bad Iris Cal.? We replaced the collimator but still same error.

Report Post
Mon Nov 18 2019Reply from C12Guy
C12Guy

The collimator iris calibration data contained in the EEPROM on the Image Function PCB is not valid. Refer to the Image System and Image System Calibration sections.

Bad Iris Cal
During boot-up, the collimator iris is driven to the maximum and minimum sizes and the
potentiometer values are compared with those stored in the EEPROM. This error occurs
when the values from current behavior do not match those saved.

 