Display N2O and anaesthetic agent

Hi I have a Cardiocap/5 which won't display the N2O and anaesthetic agent on the main screen - I know that the machine is able to measure them as somewhere deep in the service menu there was a whole table of all the parameters it was measuring (including N2O and AA). I've had to replace the SRAM battery - have I forgotten to ask for this parameter somewhere? Thanks

   RE: Telemetry-Central Station Report Post
Mon Dec 02 2019Reply from ReNew14
avatar placemark

After you replace SRAM, usually you have to go into service mode and you can tell the unit what all it has installed/ configured.

it has you pick if it has gas and then theirs sets that will start adding parameters. 

I can walk you through it if needed.

844-425-0987 and ask for Neil 
 


 

 


Report Post
Tue Mar 03 2020Reply from BrentS
BrentS

After replacing the SRAM battery the unit will lose all previous settings. You will need to correctly set the monitors "frame type" in the service menu before you can display any gases. 


Brent SilviusAnesthesia TechnicianBiomedical Sales and ServicesMercuryMed.com