Fri Dec 06 2019

Error 7.3

I need some advice on the error 7.3, DI cooling system. The laser is not warming up after a move.

   RE: Aesthetic Laser Report Post
Sat Dec 07 2019Reply from karpfen
avatar placemark

7.3
DI pressure switch does not change when power turned on.
OR
DI pump is not ON or DI pressure switch not actuated.

Inbox me and I get get u the manual


Report Post
Sun Dec 08 2019Reply from DaveSA
avatar placemark

Karpfen, can you email me the service manual instead to davegomex@protonmail.com. ?


Report Post
Sun Dec 08 2019Reply from Erik J
avatar placemark

Karpfen, can I get a copy of the service manual as well? 


Report Post
Mon Dec 09 2019Reply from TLM
avatar placemark

water pump does not work, flow switch malfunction or the water flow is blocked 


Report Post
Mon Dec 09 2019Reply from MedWrench Mandy
MedWrench Mandy

If you would like to share with the MedWrench community, please email your document to mandy@medwrench.com and we will upload it to the product page on your behalf.


Mandy Hernandez | Digital Marketing Coordinator
email me: mandy@medwrench.com
Connect with MedWrench:
Facebook
Twitter
LinkedIN

Report Post
Sat Jul 02 2022Reply from Eng Abdulrahman
avatar placemark

Kindly send me the service manual for mgy machine 

Thanks 

Bme.abdul@gmail.com 


Report Post
Mon Jul 11 2022Reply from Emily Vitkovitsky
Emily Vitkovitsky

If you would like to share with the MedWrench community, please click HERE and we will upload it to the product page on your behalf


Report Post
Mon Jul 11 2022Reply from Emily Vitkovitsky
Emily Vitkovitsky

If you would like to share with the MedWrench community, please click HERE and we will upload it to the product page on your behalf


Report Post

This site uses cookies. By continuing to browse the site you are agreeing to our use of cookies. Please review our Privacy Policy for more details.
I Agree