Err0001

Hello I have a device siemens-yasio max malfunctioning .where it appear /Err0001/ intermittently. can anyone help me

   RE: DR Systems Report Post
Mon Dec 09 2019Reply from ceshop
avatar placemark

this is normally due to the +5vdc on the CPU board being low, reseat the connector at the power supply to the cpu

 

Err001: No communication between control console and generator

Description:·  Control console could not establish a connection to the generator

Possible Causes of Error:

D300 processor (MCB2) or D500 not running.

W515 console cable defective

D501 defective.

Corrective Measures:

Check the LEDs and 7-segment displays on the D500:

No LED on: D500 power supply -> check 5 Volt

Check the D501 "F1" fuses, check power supply cable connection to D501 (W503) and D501 to D500 (W500).

Check D500 ribbon cable to D501 (W509).

Check D500 cable to D501 (W510)

Check console cables.

Replace the D500 / D501 / D300 / D502 (dongle).