Err0001
RE: DR Systems Report Post
this is normally due to the +5vdc on the CPU board being low, reseat the connector at the power supply to the cpu
Err001: No communication between control console and generator
Description:· Control console could not establish a connection to the generator
Possible Causes of Error:
D300 processor (MCB2) or D500 not running.
W515 console cable defective
D501 defective.
Corrective Measures:
Check the LEDs and 7-segment displays on the D500:
No LED on: D500 power supply -> check 5 Volt
Check the D501 "F1" fuses, check power supply cable connection to D501 (W503) and D501 to D500 (W500).
Check D500 ribbon cable to D501 (W509).
Check D500 cable to D501 (W510)
Check console cables.
Replace the D500 / D501 / D300 / D502 (dongle).