Tue Dec 10 2019

Error 4

The C-Arm is showing Error 4 on switching on and from the control Back panel at the GEN & ERROR PROCESSING board the MMPNR Led is illuminated, from the service manual am told to run the MMP tst, how can I run that test in order to get away with that error. because it doesn't operate when any error is still on. Thanks.


   RE: Mobile C-Arm Report Post
Tue Dec 10 2019Reply from ceshop
avatar placemark

is it an alarm or warnign error 4? if it is an alarm check the mains unit and fuses


Report Post
Thu Dec 12 2019Reply from KASAMBULA TONNY
avatar placemark

Dear Ceshop its a warning error 4


Report Post
Thu Dec 12 2019Reply from ceshop
avatar placemark

check the mbc power supplies.

do you have an error on the MMP?


Report Post
Mon Dec 16 2019Reply from KASAMBULA TONNY
avatar placemark

Thanks Ceshop I have AN ERROR on the MMP.

 


Report Post
Mon Feb 10 2020Reply from brunorafag
avatar placemark

Good morning Kasambula,

I have the same problem in C-arm BV29, the image no appers in the monitors and shows erro 4 on the display.

 

Did you manage to solve this problem ? Can I help me ?

Thank you in adviced!


Report Post
Wed Nov 11 2020Reply from KASAMBULA TONNY
avatar placemark

Good morning Brunorafag,

I have not solved the problem yet, but i donot know if for you you did?

if so how?

Kind regards.


Report Post