Drive error 905
RE: Slide Stainer/Prep Report Post
Error Code 905 File System Error (Drive B)
Àz An error occurred in the compact flash memory card (CF card) loaded in the CF Drive B.
A. Possible Cause
1. Any of control files necessary to operate the instrument does not exist in the CF card loaded in the CF
Drive B.
B. Tools and devices needed
* A personal computer that is able to read a CF card
C. Check and Remedy
Possible cause #1:
Initiate the maintenance program to access the “3. Program & Data File” screen from the main menu.
From that screen, execute “Data File Back up” to make a backup of the control files necessary to
operate the instrument in the CF card loaded in the CF Drive B.
Check contents of the backup to locate the missing control file (as for the control file names, refer to
the paragraph 2-20 “Control Files”). If the file name is identified, add that control file to the CF card
in the CF Drive B.
Because the folders to be added differ by control file, contact the manufacturer for more detailed
information.