Wed Jan 08 2020

Error 3

Does anyone know what would cause/ what the fix for an error 3 message is? I see that none of these have replies yet, so if i discover this myself I'll add it to the tread myself. All that being said, if you have an answer please educate me.

   RE: Urine Analyzers Report Post
Thu Jan 09 2020Reply from jroberts
jroberts

Hey Steven,

Per the Owner's Manual: Error 3: The table is not present or is notpushed in far enough to allow movement.Push the table in about halfway.

Could be an open/defective table detect switch, may need to remove the faceplate/top cover to investigate further though!


-J
Report Post
Wed Jan 15 2020Reply from Steven Bradley
avatar placemark

Turns out the user didn't lnow what they were doing. Drove all the way out to the clinic only to find they had gotten in contact with the Lab and were informed of the clinics mistake.  Yes the clinic was loading the machine incorectly.


Report Post
Fri Jan 17 2020Reply from jroberts
jroberts

a win's a win - sorry you had to drive out though!

-J


-J
Report Post