Thu Jan 09 2020

Quinton ett software problem

Hello friends, i have a problem in Quinton ett software (Q-stress software version 3.51). when everything ready and start the patient ,during this time system software hung. i was reload the system software for this problem, but system two days working fine and still have same the above problem. kindly guide me

   RE: Treadmill Report Post
Fri Jan 10 2020Reply from MedWrench Mandy
MedWrench Mandy

Although it is your choice to go offline with your discussion, we are sure that other MedWrench users would appreciate if you kept your discussion within the thread. This will help to keep a permanent record of the issue you are facing and how it was solved to help others in the community that may be facing a similar challenge.  

We appreciate your participation and hope that the MedWrench forums will help you to solve your medical equipment issue. 


Mandy Hernandez | Digital Marketing Coordinator
email me: mandy@medwrench.com
Connect with MedWrench:
Facebook
Twitter
LinkedIN

Report Post
Tue Jan 14 2020Reply from Connect Wright
avatar placemark

Check Atlantis DB 1,2 & 3 if any are above 75% you will need to clean DB.