Sun Jan 19 2020

Temp probe issue

On the lower chamber, I am getting extremely out of range readings that do not match actual or set temp. Could this be the temp probe? if so, where is it located and how do i get to it to replace it? thanks.

   RE: Warming Cabinet Report Post
Mon Jan 20 2020Reply from Chris35
avatar placemark

Item #3 in the drawing below. You may be interested in item #39, enhancement kit for both probes. Could also be the control board but most likely a bad connection between the probe and the control.

Exploded view


Report Post
Wed Jan 22 2020Reply from Matt D.
Matt D.

Good information! Thanks for sharing!


Matt D.

Report Post
Wed Jan 22 2020Reply from Chris35
avatar placemark

I should add...

I was refering to a Gen 5 warming cabinet with the small black temp adjustment knob and sliding control cover panel. Since the 2 pictures shown here (above left) are for 2 different models. If yours has a blister touch pad for temp control, I would still verify connections, but it is most likely a defective control board.