No air flow

Hi all, I have start up the long-unused RapidLab 348ex but the gas flow is blocked. (Observed no bubbles coming out of the probe when calibrating gas, resulting in slope calibration failure.) Please could you help me with this issue? Thank you very much.

   RE: Blood Gas/pH Analyzer Report Post
Fri May 29 2020Reply from ELECTROMEDMDP
avatar placemark

22/5000

did you solve the problem?


Report Post
Fri May 29 2020Reply from Mc Grath Carlos
avatar placemark

Please check the gas containers they must be new 

Check replaco pump tubing 

Check the valve that commands the gas flow to the system 

Fell free contact me 


Report Post