Hi all,
I have start up the long-unused RapidLab 348ex but the gas flow is blocked. (Observed no bubbles coming out of the probe when calibrating gas, resulting in slope calibration failure.)
Please could you help me with this issue?
Thank you very much.
Wed Jan 22 2020
No air flow
Fri May 29 2020Reply from ELECTROMEDMDP
Fri May 29 2020Reply from Mc Grath Carlos
Please check the gas containers they must be new
Check replaco pump tubing
Check the valve that commands the gas flow to the system
Fell free contact me