Calibration Error

We have been calibrating our Sentry 1400 pumps from Tridien Medical for some time. Now every time we calibrate a pump (using the same method we always have) we keep getting Calibration Errors on the last 3 pumps we have calibrated. We have tried redoing the process like the manufacturer suggests when getting an error, but still the calibration error remains. Looking to see if anyone has had this or a similar issue with this product or similar products. It is a Sentry 1400 from Tridien Medical. Low-airLoss, alternating pressure mattress system. Thanks.

   RE: Mattress Report Post
Mon Sep 14 2020Reply from Service guy
avatar placemark

We have the same pumps whats the procedure for calibrating pump.  Never been able to erase calibration error message


Report Post
Mon Jun 07 2021Reply from
avatar placemark

Does the calibration procedure. Require applying a known pressure to the transducer?


Report Post
Tue Jun 08 2021Reply from PatOps
avatar placemark

The procedure requires 60 mmHg of pressure to calibrate the Sentry 1400 control unit