VOLUSON 730 PRO V DOES NOT TURN ON
RE: Ultrasound Systems Report Post
One of these companies may be able to help you:
Mandy Hernandez | Digital Marketing Coordinator
email me: mandy@medwrench.com
Connect with MedWrench:
Report Post
is the DC Voltages OK? (+5V,+15V,-15V,+12V,+3,3V) if not then it could be the CPP Board. if the Voltages are OK try remove CKV-Board and connect the monitor directly on PC VGA-Card then power on the system. if still won't turn on trouble the PC or Kontron Board.
Report Post
hello thanks for answering, the CPP card does not give the tensions, the LEDs that indicate if there is voltage (+ 5V, + 15V, -15V, + 12V, + 3.3V), they do not light up, of the 6 LEDs it only turns on the led that of the fan voltage. My question is: is it necessary for the CPP to detect load pars to stay on?
Thanks
Report Post
I would recommend checking the fuses on the CPN.
Depending on which CPN is in the machine there could be more or less.
-Conquest Imaging Tech Support