my amx 4 plus presents error 45 c generator failure
Sun Feb 23 2020
AMX error 45c
Sun Feb 23 2020Reply from oldxrayguy
error 45C
1. Check For 80 VAC At CR4 Black Leads On AMX1A4 Panel With
PREP Pressed
2. Check For About 75 VDC Across AMX1A4 C2/C12 With PREP
Pressed
3. Check For Correct Polarity Of CR1 & CR2 (CR7/CR13)
On AMX1A4 Panel
4. Bad 1KHZ Board
5. Are The Tap Relays Energizing?
6. Is “Stop Cap” At J4-11 (1 Khz Board)
Approximately 80 VDC?
ERROR