Advisory 461- infiniti back up power service needed, contact alcon technical services. How do I solve it?  should I change the battery? Thank you for your help.


   RE: Phacoemulsification Unit Report Post
Thu Feb 27 2020Reply from John Weymouth
John Weymouth

Hello Marcelo!

Make sure the power is off for the system and it is unplugged.

The 12-volt battery that is located behind the metal panel inside where the foot pedal is located.  You will need to move the levers on the side of the drawer to get enough room to remove the plate with the two capture screws.

The batter has a thermal sensor that needs to be on the battery and under the foam tape.  The battery is held down by a bracket with one screw.

After replacing the battery, you will need to make sure no wires are pinched and the metal plate is secured.

If you need further assistance, we will have an online PM course for the INFINITI system soon.  Let me know if that is of interest. 

John
 