no signal error

I am getting a self diagnostics check signal cable no signal error on a Siemens Acuson X150. I have checked the cables to the unit. The only one disconnected the staff report not using and plugging it in does not clear the error. I have rebooted the unit and still have the same error. Siemens is not providing tech support on this unit. Is there a service manual available? Does anyone know how to clear this error?

   RE: Ultrasound Systems Report Post
Fri Mar 06 2020Reply from MedWrench Mandy
MedWrench Mandy

Mandy Hernandez | Digital Marketing Coordinator
email me: mandy@medwrench.com
Connect with MedWrench:
Facebook
Twitter
LinkedIN

Report Post
Fri Mar 06 2020Reply from Mike Davis
avatar placemark

The issue you are having is with the BE board and RM board. These boards are the main computer of the system and are failing to start. You will want to repalce both parts.

For serial # less then 307998 use 10131804 and 10349420

For serial # greater then 307999 use 10349464 and 10349439


Michael Davis
Technical Support Specialist
Conquest Imaging
Report Post
Fri Mar 06 2020Reply from R J
R J

It could be power supply issue. Verify AC / DC Power Supply LEDs located on the back of the system. If DCDC power supply is bad then DC LEDs is solid amber (yellow orange) the system might have over current or over voltage or over temp. Same as the AC LEDs.

info@emedicalsol.com

EXCLUSIVE MEDICAL SOLUTION


Report Post
Fri Mar 06 2020Reply from Christina B
avatar placemark

Thank you. 


Report Post
Sun Nov 22 2020Reply from FrkGngor
avatar placemark

 

I am having the same mistake. ''self-diagnosis, check signal cable '' There is no problem for DC / DC power and AC / DC power. "Self-diagnostic control cable no signal" gives this error. Can your Help Please.