hello, i need service manual of Flow-i C20 . it has power switch probably broken . say error T61 power switch .
someone can help me?
thx in advance
Mon Mar 09 2020
t61 error
Mon Mar 09 2020Reply from Clint Ech
I haven't seen a TE61 error but have has a power issue. One question - does the fan below the patient cassette run when you press the power button? If the sensor in the fan has failed, the machine won't start up.
Mon Mar 09 2020Reply from fabiomirko
hello. if you press power switch one time machine start and after few
seconds goes off.. if you keep press the switch for 10 -12 second ,machine starts without problems... it pass all test and all is ok but on lcd say " power switch error"
If the fan is running then it may be a switch issue.
Next i would try without the switch in the circuit but be careful when bypassing the switch as you will be joining live wires once switched on at the plug.