t61 error

hello, i need service manual of Flow-i C20 . it has power switch probably broken . say error T61 power switch . someone can help me? thx in advance

   RE: Anesthesia Gas Machine Report Post
Mon Mar 09 2020Reply from Clint Ech
avatar placemark

I haven't seen a TE61 error but have has a power issue. One question - does the fan below the patient cassette run when you press the power button? If the sensor in the fan has failed, the machine won't start up.


Report Post
Mon Mar 09 2020Reply from fabiomirko
avatar placemark

hello. if you press power switch one time machine start and after few

seconds goes off.. if you keep press the switch  for 10 -12 second ,machine starts without problems... it pass all test and all is  ok but on lcd say " power switch error"


Report Post
Mon Mar 09 2020Reply from Clint Ech
avatar placemark

If the fan is running then it may be a switch issue.

Next i would try without the switch in the circuit but be careful when bypassing the switch as you will be joining live wires once switched on at the plug.