Probe port problem
RE: Ultrasound Systems Report Post
Also, if someone has a service manual that has the proper disassembly of the FRU's or individual boards on the machine I would appreciate it.
Thanks...
Report Post
One of these companies may be able to help you:
Mandy Hernandez | Digital Marketing Coordinator
email me: mandy@medwrench.com
Connect with MedWrench:
Report Post
The issue is with the port itself. This is not a field repairable part. You will want to replace the entire Probe Interface Board. Part # 5366065-3
Michael Davis
Technical Support Specialist
Conquest Imaging
Report Post
Howdy,
The first port is for 'parking' a probe. There are only 3 active connectors on the Probe Interface Board of a Voluson S6.
Best regards,
Jim Carr
Director of Service & International Operations
AUS
Toll Free (USA, Puerto Rico and Canada): 866-620-2831
USA office: +1 918-628-2831
Email: Service@advancedultrasound.com and Sales@advancedultrasound.com
Report Post
Hi Guys!
Thanks Mike Davis for the info. I have already instructed the end-users not to use the said probe to avoid further damage to the port itself.
Thanks also Jim Carr for the advice although in this case, the fourth probe port is the parking slot not the first one.
Best Regards,
Report Post
Also, could you send me a copy of service manual which has the proper disassembly of the probe PCB? would appreciate it thanks
Report Post