Got e13 error on Smith& nephew 500xl. Replaced bulb, but error still appears. Temp sensors are ok. Can anyone assist?
Fri Mar 20 2020
E13 error
Mon Mar 23 2020Reply from MedWrench Mandy

iMed Biomedical may be able to help you.
Mandy Hernandez | Digital Marketing Coordinator
email me: mandy@medwrench.com
Connect with MedWrench:
Report Post
Mon Nov 30 2020Reply from BiomedBailie
Having the same issue with two units producing E13. I have discovered that the same failure on both units is R10 of the power supply board. Anyone know the resistor value? Likely to happen again and suprised there has not been a recall on these... Anyone have a solution for us!?
Thanks in advance,
Biomed Bailie