Tue Apr 28 2020

Megafuge 8

I have a Megafuge 8 that is having an issue with the motor not running we replaced the main board and still not running. Tech support stated the board could be bad and sent us a new one but still the same issue not giving any error messages. My next guess is that it maybe a latch assembly any one have this problem .

   RE: Centrifuge Report Post
Tue Apr 28 2020Reply from khush.chera_7817
avatar placemark

Dear Hector,

It is more likely "LID LATCH ASSEMBLY" has malfunctioned.

Usually 1 X Zippy Microswitch (there are 5 microswitches in the assy), located where the Lid Latch engages with the LID LOCKING ASSY, IT IS A LONG ARM MICROSWITCH has broken contact.

When you close the Lid, Lid Latch Motor spins, but due to broken contaact it does not able to send a message to main cpu that it is ready for run

Solution: Replace both side LID LATCH ASSEMBLIES, it will be expensive or if you are willing to strip it further then replace this faulty  ZIPPY MICROSWITCH (Normally Open and also waterproof)

 


Report Post
Tue May 12 2020Reply from Ben C
Ben C

Ozark may be able to assist. 


Find me on FacebookLinkedIn, YouTube or Twitter 
 

Report Post

This site uses cookies. By continuing to browse the site you are agreeing to our use of cookies. Please review our Privacy Policy for more details.
I Agree