I have a Megafuge 8 that is having an issue with the motor not running we replaced the main board and still not running. Tech support stated the board could be bad and sent us a new one but still the same issue not giving any error messages. My next guess is that it maybe a latch assembly any one have this problem .
Tue Apr 28 2020
Megafuge 8
Tue Apr 28 2020Reply from khush.chera_7817
Dear Hector,
It is more likely "LID LATCH ASSEMBLY" has malfunctioned.
Usually 1 X Zippy Microswitch (there are 5 microswitches in the assy), located where the Lid Latch engages with the LID LOCKING ASSY, IT IS A LONG ARM MICROSWITCH has broken contact.
When you close the Lid, Lid Latch Motor spins, but due to broken contaact it does not able to send a message to main cpu that it is ready for run
Solution: Replace both side LID LATCH ASSEMBLIES, it will be expensive or if you are willing to strip it further then replace this faulty ZIPPY MICROSWITCH (Normally Open and also waterproof)