Thu Apr 30 2020

Stat Spin 4 rocks /vibrates

Our Stat Spin rocks / vibrates upon start up and at the end of cycle when balanced. I don't have another rotor to test with and was wondering if this was the issue ?

   RE: Centrifuge Report Post
Wed May 06 2020Reply from rstopczynski_5658
avatar placemark

Take the rotor out and look on the bottom for cracks. I just had one do this and that was the problem.


Report Post
Wed May 06 2020Reply from TRS
avatar placemark

Most likely has a cracked rotor.  I would recommend inspection of the motor mount as well.  Ozark Biomedical part number is 10238 for rotor.  May want to check imbalance switch adjustment additionally.


Report Post