service light
RE: Defibrillator Report Post
Have you replaced the battery connectors? (small bananna plugs under the battery end). These will eventually corrode up and loose connectivity. The PM process calls you to replace these every couple or so years.. I have seen issues like yours if you do not replace them regularly.
mrstan
mrstan.
Report Post
Sometimes cleaning the device internally, clears the wrench as well as what Stan mentioned.
Report Post
Stan has it correct.
Regards, Michel.