Philips C arm BV 25 1993 is not giving an image?

Dears gREETINGS We are using C arm bv 25 1993, we are facing a problem when making an exposure, no image display on the monitor - the x ray is ok, by using old film - no errors display - The radiation LED is turns ON while exposure please help, what possible cause for this thanks

   RE: Mobile C-Arm Report Post
Tue May 26 2020Reply from MM - Porto
avatar placemark

BV25 only gives image on Fluoroscopy, not in radiography

If you have X-Rays and no image in Fluoroscopy:  Image Intensifier, TV Chain,... 

I had once a problem with Image Intensifier: not ready. The cause was the Power Supply of II defective (no 30KV to II to function). This power suply is located on II flat part, you have to disassembly the C from the II, 4 screws. Be careful.