Tue Jun 30 2020
COUCH´S VERTICAL MOVEMENT FAULT
The system won´t move the couch in the vertical direction. If I press the buttom switch on the gantry, the leds goes off for a while, but the couch doesn´t move. However, when I select the service switch, the couch can be move up and down. I already try to flash the CPM, check interlocks and voltages.
RE: CT Scanner Report Post
Tue Jun 30 2020Reply from Ralph!
Tue Jun 30 2020Reply from Wilver
Wed Jul 01 2020Reply from Ralph!
I have found in the past vertical table issues were resolved by replacing the Inverter Panel assembly part #453567484751
Ralph
Wed Jul 01 2020Reply from Wilver
Ok. Thanks.
I was thinking on the inverter too; however, I wonder why it works in service mode. If I turn the key in the gantry to service and put the service switch to on on the couch, I can move with the push buttoms the couch UP and DOWN.
Regards,
Wilver
Thu Jul 02 2020Reply from Ralph!
You are correct. You are in service it bypasses all the normal operation. It is just the switches and motor. So if in an failed situation you can get the patient out safely.
Thanks
Ralph