When patient monitors displays loss of memory which particular component is dead or weak ?

Report Post
2 days agoReply from MedWrench Kari
MedWrench Kari

What is the name of the manufacturer and model of the equipment you are working on? 


Kari Thompson | Data Entry Coordinator                     email me: kari@medwrench.com
Report Post
2 days agoReply from mccariel
avatar placemark

GE Dinamap pro 1000 as well GE dash 2500. Same loss of memory issue

 


Report Post
a day agoReply from S. Rooks
avatar placemark

Hello Mccariel,

The internal, 3.6 V NiMH battery on the main board acts as an alarm backup and maintains the nonvolatile RAM memory when the Monitor is off or away from AC mains. Normally, the Electronically Erasable Programmable Read Only Memory (EEPROM) is a chip that is used to store the calibration and other "setting" variables that have to be maintained in the event of a complete power failure. It is possible that a backup battery replacement is required.


Biomedical Equipment Technician with Southeastern Biomedical 1-888-310-7322
Report Post
a day agoReply from mccariel
avatar placemark

Thanks l ll change this battery and see what happens .Many regards to the great members of this forum always out there to help

 