Hello everyone ,I get a gray imagen after radiation .
Apparently ,the system is working because no error message and the Kv and MA values appears at the control panel.
Thanks for your help-
Sat Jul 04 2020
GRAY IMAGE AFTER RADIATION
Sun Jul 05 2020Reply from ffriedmann
While you state that you have kV and mA values on the panel, you don't give the values. Or what you had in the field. Put about 2.1 mm of copper in the field and see what the values are. Should be in the mid 70kV range or so, check the test certificate in the system book for the values. Need to know these values to further assist.
What does a stored image already on the system look like? Do you get any reading with a dose meter?
Report Post
Mon Feb 05 2024Reply from Mahran97
Hello,
I have the same problem did you manage to fix it,
Your feedback is highly appreciated