I have a problem with Philips - Brilliance 6 - shoots once and stops and in the majority of cases it does not shot.

Hi all, i am a Doctor working in the war zone of Idlib. we dont have any biomedical engineers here to help us, so it is us medics trying to repair devices. Right now we are having problems with one of the only CT scanners we have working here. it is not warming up and the device is giving the following errrors in the log. can anyone help. S_COLLISION_BAD_DTH_PARAM S_DTH_UNKNOWN_USER  - 

S_GAFSTARTUPTASK_DTH_BAD_PARAM -

S_MASTER_SEQUENCER_EXIT_STATE_ERR  -

S_MASTER_SEQUENCER_NO_XRAY  -

S_RHOST_HVG_HSS_GENERATOR_NON_ARC_FAULT  -

S_RMC_DTH_BAD_PARAM S_RMC_MP_DTH_BAD_PARAM S_RMC_SB_FAULT -

S_ROTOR_DTH_BAD_PARAM - 

 


   RE: CT Scanner Report Post
Wed Jul 08 2020Reply from Ben C
Ben C

One of these companies may be able to assist:

Richardson Healthcare

RSTI Imaging

Tri-Imaging

TTG Imaging


Find me on FacebookLinkedIn, YouTube or Twitter 
 

Report Post
Mon Jul 13 2020Reply from FEIJING MEDICAL
avatar placemark

1. Check the external power supply and the power supply voltage in the cabinet. 2. X-RAY bulb heating and preheating abnormalities are mostly related to the filament power supply, the vacuum degree of the bulb, and the filament resistance. 3. The exposure times of the bulb exceed 300,000 seconds. Replace the tube. If you need help, please email me, radio05@sina.com Mr Ding