Arcadis Varic Gen2, a lot of errors.

Hi everyone Please help, I found errors as followed

14 _ASP_BCC Unable to noitfy other comp...

247_ASP_AFE Reset the current application...

92 _ASP_ASM The synchronization of the...

4138_ASP_MSY Image System mode change ti...

98 _ASP_ASM Error recovery failed. Pres...

116_ASP_ASM The requested action could...

110 _ASP_ASM Change Mode failed...

44 _CAP_PR Network is Down.

3 _CAP_FL An unknown error has occurr...

117_ASP_ASM Cannot connect to database.

327_ASP_ABE Examination Taskcard: Unable...

190_ASP_ABE A general problem with the...

50081_ASPASR Creation of Radiation Sumar...

50082_ASP_ASR Saving of the Radiation Sum...

103 _ASP_ASM Patent registration failed.

69 _ASP_ASM The requested action could...

 

These errors happen round and round. Please advise.

Regards,

Chao.

 


   RE: Mobile C-Arm Report Post
Fri Jul 17 2020Reply from Tomelgy
avatar placemark

Do these errors also appear without the main system connected?

Can you see both task cards on both monitors?

Have you tried looking in explorer to see if all 3 HDD are present/visible?

You seem to have some errors only releated to sending images to PACS.

When do these errors appear, at power-on or when registering a patient?

Can you see and open previous patient images?

Tom


Report Post
Sun Jul 19 2020Reply from Chaovalit
avatar placemark

Hi Mr.Tom,

This equipment still working.

Customer found problem when they tried to transfer data to PACS that some pictures was lost.

Is It possible harddisk mal-function?  Please advise.

Regards,

Chao.


Report Post
Sun Jul 19 2020Reply from Tomelgy
avatar placemark

Yes of course the HDD could be giving issues, you could try via windows run check disk, you can also run defrag which can help speed things up.

Normally it would be unusual to loose individual images, or whole patients. It could be the images was not saved by customer. Or if patient lost then look towards the browser menu for filter, you can set exam date, entry date, name etc. Make sure all filters are off then look for patient again.

The Arcadis does not like the database being accessed for sending to PACS while trying to aquire images, you can only do one or the other not both.

If the images going to PACS were lost, then check to make sure there is NOT a duplicate IP being used.

Tom


Report Post
Sun Jul 19 2020Reply from Chaovalit
avatar placemark

Hi Mr.Tom,

Thanks for your kindly support.

Can I replace all harddisk with SSD?

Regards,

Chao.


Report Post
Mon Jul 20 2020Reply from Tomelgy
avatar placemark

There is no official instruction to use SSD.

Of course it may work, but there are many different SSD available, I would keep to the existing HDD model.

If you use SSD I would image you will need to reload software, which means you will need a license key that will allow administrator access!

The problem does not suggest a HDD, you have far to many different, unrelated errors, There is more going-on than being advised.

A copy of the developer logs, time/date of problem would help. To obtain the developer log, press Ctr+Alt+S keys, after a few seconds a windows will appear saving the developer log.

To retrieve the developer log, go to C drive, service & look for the 'extract' folder. in there you should see many logs in date order the top log should be the developer log with system SN.

Once you have the developer log copy it to a USB drive, rename the extension with (for example) 12000developer etc. to 12000developer.zip.

Then extract the zip file & look through the various logs.

Once you have the log file, send me your email.

Tom


Report Post
Wed Jul 22 2020Reply from Chaovalit
avatar placemark

My email address is  guai38@yahoo.com

 

Regards,

Chao


Report Post
Wed Aug 19 2020Reply from Chaovalit
avatar placemark

Hi Mr.Tom,
I would like to inform you, after I did as your advice from that date.

Today I ask the customer which any problem, They said no problem up to now.

Thanks for your kindly support.

Regards,

Chao.


Report Post
Wed Aug 19 2020Reply from Tomelgy
avatar placemark

Hi Chao

Great to here, thanks for the feedback.

Tom


Report Post
Mon Apr 05 2021Reply from Amin Z
avatar placemark

Hello,

I have got two error from Siemens Arcadis Varic:

ASP_MBE_W_17

ASP_MBE_E_70

ASP_BIC 1001

 

Could you please advise?

 

 

Best regards,

Amin


Report Post
Mon Apr 05 2021Reply from Tomelgy
avatar placemark

Amin

MBE_W17: Forced shutdown, jobs pending. (Possible CD burning)

MBE_E70: No information. Is this correct error?

BIC_1001: During startup, hardware error, (Possible CIPP board or BNC disconencted).

Do the errors appear with just the Cart (Main system disconnected), No network connected, no additional USB or HDD connected.

Check the X10 connector pins for damage, a couple of pins are shorter (normal). 

Tom