Ventilator Inoperative

Good Day! Hoping someone can help us with our problem. We have 2 units of PB840 and they're not working anymore. Unfortunately, we don't have a BioMed. Can someone please help us how to troubleshoot the ventilator. It appears "Ventilator inoperative". What does it mean? How to fix it? Thank you!

Report Post
Fri Jul 17 2020Reply from T-Lobato
T-Lobato

Here's what the manual says, "In some situations, the ventilator declares a ventilator inoperative condition. A hardware failure or critical software error that could compromise safe ventilation triggers this condition. When this occurs, the ventilator inoperative indicator lights and the ventilator enters the safety valve open (SVO) state. To correct a ventilator inoperative condition, the ventilator must be turned off, then powered on again; at power on, a qualified service technician must run extended self test (EST). The ventilator must pass EST before normal ventilation can resume."

Not knowing where you're located, I would recommend calling a local hospital and talking with the Biomed department for possible service providers.  You could also call Medtronic and ask them for assistance.

Good luck!


TJ Lobato | MedWrench Guru
Report Post
Sat Jul 18 2020Reply from NarsCCU
avatar placemark

Thank you so much for your kind reply.

Unfortunately, our ventilator failed the EST more than 3 times. It used to work after a succesful EST whenever we use it. However this time, every after EST the ventilator flashes "ventilator inoperative". 

I really appreciate your help, didn't expect you'd reply. 

 

Godbless!


Isabela

Report Post
Sat Jul 18 2020Reply from jrpiceno
avatar placemark

Hi,

Are you using the correct test circuit? That is very important to have. I would take note of each test that failed in the EST and start there. If you can start the EST, I would say that is a good indication the communication between the GUI and BDU is ok. Has your unit been sitting for a while or in use? These have also reached the end of manufacturer support. You can still get parts from vendors like PartsSource if needed.

Hope this helps. 
 

JP


Report Post
Sat Jul 18 2020Reply from NarsCCU
avatar placemark

Hello!

Thank you for replying. We're using Hudson ventilator circuit ever since. Our Ventilators are stocked for more than a month now. We tried everything we can--EST everyday, but the same result and it also shows OVERRIDEN. If you have some ideas or advices on how to troubleshoot our ventilators, we're willing to try it. 

Thank you and God bless you!


Isabela

Report Post
Tue Jul 21 2020Reply from Amscopolaris
Amscopolaris

Flex tube, 21.0 in. (53.4 cm) (gold standard tube) (2 required) 4-018506-00 EST, performance verification

if there is no tube in this feature, est test gives error 


Report Post
Tue Jul 21 2020Reply from CMB7
CMB7

Isabela
 

I am a RRT - you can email me with the following info please:

Please reply with all the information on the unit's ID tag:

  • Serial number -

  • Model number -

  • Year of manufacture -

PLUS

  • Software updates? Yes___ No___

  • Nature of the problems:

 

Regards, Michel.

C. Michel Bessette, RRT (Canada #72-659)

Servicios Tecnicos CMB7

535 Canteval,

Ottawa, Canada

serviciostecnicoscmb7@gmail.com

Mobile: 6138185321

 

Bio-Medical Technological Services: Compressed Gas Systems Specialist

Anaesthesia & Hyperbaric Chamber Technologist

Services in: English, Français & Español

 


Bio-Medical Technological Services: Compressed Gas Systems Specialist
Hyperbaric Chamber Technologist
Services in: English, Français & Español
Report Post
Thu Jul 23 2020Reply from NarsCCU
avatar placemark

Good Day Sir! I already sent you an email. Thank you!


Isabela

Report Post
Fri Jul 24 2020Reply from NarsCCU
avatar placemark

Hello Sir! Do you have another email? It says that the email address is not found. Thank you!


Isabela

Report Post
Mon Jul 27 2020Reply from Manford
avatar placemark

Hi Isabela,

Have you solved the PB840 problem on "ventilator inoperative"?

Doing an EST on 840 vent needs a "gold tube" without a inlet filter.  Some biomed make this mistake putting a filter on and not work.  Plus you must be trained on this vent.  Some hospital RRT run EST and some hospitals only allow biomed to do.

If you run a PVT, you need a PTS-2000 (calibration date not expired) plus two gold tubes plus a proper vent circuit and some accessory like 840vts vent software and plug...    You cannot do PVT without factory trained and equipment.  Ask Medtronic (previous Covidien) for help.

You question might be a very simple issue as I saw many times came in with this message.

Sincerely

MK


Report Post
20 hours agoReply from CMB7
CMB7

Isabela, here is my contact info:

C. Michel Bessette, RRT (Canada #72-659)

Servicios Tecnicos CMB7

535 Canteval,

Ottawa, Canada

serviciostecnicoscmb7@gmail.com

Mobile: 6138185321


Bio-Medical Technological Services: Compressed Gas Systems Specialist
Hyperbaric Chamber Technologist
Services in: English, Français & Español
Report Post

This site uses cookies. By continuing to browse the site you are agreeing to our use of cookies. Please review our Privacy Policy for more details.
I Agree