I have a corometrics 170 series(172)..no leds lit on display, power and recorder leds are lit up and can hear hear beat..can not turn unit off wo unplugging, cannot turn volume down, recorder will not feed and nothing lit on display..a lot of issues at once..

   RE: System Report Post
Mon Jul 20 2020Reply from Ali Youssef
Ali Youssef

That could be a bad main board. If you send it to USOC we can check it for you. 
thanks 


Report Post
Mon Jul 20 2020Reply from Ben C
Ben C

USOC may be able to assist. 


Find me on FacebookLinkedIn, YouTube or Twitter 
 

Report Post
Mon Jul 20 2020Reply from Ali Youssef
Ali Youssef

I would say 80 percent main board. I have seen it a lot of times 

thanks


Report Post
Wed Sep 23 2020Reply from Cherry chu-JRPanel
Cherry chu-JRPanel

Sir,
Perhaps you can check whether the circuit of your device's membrane buttons is in normal contact.
Cheers,
Cherry Chu
International development of membrane switch and acrylic panel