Midmark 630 - how/what to test before buying used
RE: Power Report Post
The 630 is a great chair, except...for a few things.
The foot switches under and along the bottom edge of the chair are highly susceptible to liquids
the foot section is difficult to clean out.
When those switches are damaged, only the back function will work.
Also the over travel switches can be a pain in the butt.
Another major issue is the location of the front connection for a foot control. Housekeeping tends to short this out when mopping, causing loss of functions and annoying control issues.
You can still get upholstery for it, but not in all colors.
The knob on the head section is flimsy plastic, over tightening will break the plastic, eliminating the ability to secure the head section where you want it.
Those are the only issues I've come across since the 630 was released.
You should be able to find a midmark authorized service company in your area, and parts are still available from midmark and several 3rd party sellers.
Thanks for the tips Chris. I'll definitely check all of these areas and hopefully the chairs check out. Would you recommend to stay away from 630s after a certain year or it doesn't matter?
Thank you
As long as you cover the foot section with a bag or some type of plastic if your doing anything involving liquid, the 630 is a great chair
We don't use much liquid during our procedures so we should be ok assuming the cleaning personell doesnt' make it a problem...
Saw the chairs yesterday and they checked out perfectly. They got delivered today and everything functios well. Any recommendations on of what I can do for initial maintenance check?
Thank you