Does anyone have a PM procedure and what is the frequency of PM, or are you risking out or doing AEM.
Maintenance requirement
Rudy,
I took a look at their website but could not find a service manual. The closest thing I could find to maintenance requirements is this (bold text is mine):
Controlled oxygen delivery (when required)
Oxygen from a concentrator or cylinder can be added. An in-built ultrasonic analyzer requires no calibration, service, or replacement.
https://www.fphcare.com/us/hospital/adult-respiratory/optiflow/airvo-2-system/
If it were me, I would contact the vendor and ask about preventive maintenance.
T
TJ Lobato | MedWrench Guru
Report Post
Per the Technical Manual, "Arivo 2 humidifiers do NOT require routine servicing or calibration". There is a Performance Check option, however; there's no manufacturer requirement to carry out these checks on a routine basis as stated in the Technical Manaul. This was also confirmed via email with F&P's Quality & Regulatory Specialist and we have that correspondence on file in case the no-service/PM strategy is questioned by surveyors. You may wish to do the same. In the meantime, I'll send the Technical Manual to MedWrench's Admin for posting. Hope this helps. - Cipher