Tue Aug 11 2020

warning 206 ultrasound and coagulation not available restart system ....

how can I solve this problem ;is this problem getting from software or hardware error

   RE: Phacoemulsification Unit Report Post
Tue Aug 11 2020Reply from T-Lobato
T-Lobato

Ya easy,

think these units require a service school in order to maintain them.  That makes it tough to work on, and even to get a service manual. 

You have probably already restarted the system.  If it were me I would try that, being sure to remove power for a good bit of time to allow all the components to discharge.  If that worked, in my experience those are short lived victories as the problem will usually resurface. So, I would give the unit a good physical inspection looking for anything that might be amiss, like broken connectors, loose connections, etc. I would also do an electrical safety test.  I would expect it to pass but you never know.

T.


TJ Lobato | MedWrench Guru
Report Post
Tue Aug 11 2020Reply from John Weymouth
John Weymouth

206WarningWarning xxx: Ultrasound and Coagulation not available. Contact Alcon Technical Ser- vices.Broker s/w error.
Broker bus warning.
Broker bus off.
Broker bus queue full.
Broker bus transmit busy.
Host proxy s/w error.
Host proxy timeout.
Host proxy range error.
Footswitch proxy s/w error.
Footswitch proxy timeout.
Footswitch proxy range error.
Footswitch mechanism fault.
Fluidics proxy s/w error.
Fluidics proxy timeout.
Fluidics proxy range error.
Fluidics mechanism fault.
A/D Converter s/w error.
A/D Converter data overrun.
A/D Converter no data.
Voltage diagnostic s/w error.• CAN Bus Cable
• US/COAG Controller PCB• U/S or NeoSoniX™ Tune status is “Not Tuned”. AquaL- ase® and or AquaLase® Single-Use Tune status not affected.
• Prime status is unaffected.
• Test Handpiece button in Setup screen is ghosted if current handpiece is U/S or NeoSoniX™. Test Hand- piece button in Setup is not ghosted if current handpiece is either the AquaLase® or AquaLase® Single-Use.
• FP3 in U/S and NeoSoniX™ steps is not functional (user can go to FP3 but there will be no U/S power).
• If the Coagulation button is pressed, the subsystem sta- tus dialog is displayed.


Service pecialists with Eye Surgical Equipment, ALCON AMO B&L Call us today for Service, Sales, Training, and PARTS. One-Stop Shop for Eye Surgical Equipment. John Weymouth John@medisurg.com 317-663-3158

Report Post
Tue Aug 11 2020Reply from John Weymouth
John Weymouth

It is more than likely the PCB for the COAG/US.  

I sent you the error code readout.  I have this PCB in stock.   I also offer Biomed Training on this system and in return, you get a discount on parts.

 

John Weymouth 

 


Service pecialists with Eye Surgical Equipment, ALCON AMO B&L Call us today for Service, Sales, Training, and PARTS. One-Stop Shop for Eye Surgical Equipment. John Weymouth John@medisurg.com 317-663-3158

Report Post
Wed Aug 12 2020Reply from Ya easy
avatar placemark

I changed thePCB for the COAG/US . I bought a panel from eBay that is not new The seller says it is running (but it might  also have a problem) but the error message still appears. Note that the error message does not appear permanently Sometimes they appear during work and when the device is restarted, the device returns to work normally. 

Is your board new? How much is it?


Report Post
Wed Aug 12 2020Reply from Ya easy
avatar placemark

At the start of operation of the device A message appears that can be skipped Its content is that aqualase not available


Report Post