Service Issue

In Vela Carefusion--What is the reason of High VTe than VTi? and FiO2 monitoring calibration Failure ? O2 pressure is set at 50 PSIG and new O2 sensor changed. In normal ventilation we unit is monitoring 60-65% FiO2 while set FiO2 is 100%

 

Also needs user manual and sevice manual for this Ventilator.


   RE: Ventilator Report Post
Fri Aug 21 2020Reply from Ben C
Ben C

A copy of the documents for this equipment can be found HERE.

And one of these companies may be able to assist you:

Agiliti  

Mobile Medical


Find me on FacebookLinkedIn, YouTube or Twitter 
 

Report Post
Sat Aug 22 2020Reply from CMB7
CMB7

Hello Tharu,

Everthing you need for that ventilator is here:

https://www.vyaire.com/products/vela-ventilator

Regards, Michel.

C. Michel Bessette, RRT (Canada #72-659)

Servicios Tecnicos CMB7

535 Canteval,

Ottawa, Canada

serviciostecnicoscmb7@gmail.com

Mobile: 6138185321

 

Bio-Medical Technological Services: Compressed Gas Systems Specialist

Anaesthesia & Hyperbaric Chamber Technologist

Services in: English, Français & Español


Bio-Medical Technological Services: Compressed Gas Systems Specialist Hyperbaric Chamber Technologist Services in: English, Français & Español

Report Post
Wed Dec 30 2020Reply from daniel biomedics
avatar placemark

 

Hi good day. You must check the altitude where you are located and adjust this height on the fan. This procedure is essential for o2 cell calibration. it also depends a lot on the useful life of the o2 cell. Regarding the inlet pressure, the carefusion ventilator admits from 40-85psi, being the internal regulator who limits said pressure.


Report Post