System error, please restart
RE: Ultrasound Systems Report Post
Base Software: 5135124-30
The error is kinda generic. Could be anything from improper cabling to the Operator Panel to the Main Power Supply. Retrace the last repair/PM if anything was done recently.
The link below is to GE Documentation library pick a modality and the documentation you desire and download.
https://www.gehealthcare.com/products/support-documentation-library
Report Post
Thanks for this infos.Only i did not see anything anormal on a câblage. i think i will have to réinstaller a base software.
I am first look for him.
Report Post
Hi, I usually see this error if there are corrupt files in the OS, but other things can cause it too. After checking cable connections and voltages at the BEP power supply, I'd try to start the machine and make sure all the BEP fans and TXPS fans are running (on the bottom of the main power supply you can see them with a mirror).
You could also try minimum configuration, or at least disconnect the Front End cable from the backplane inside the BEP. If you can get it to run in simulator mode, it may be that a front end card is dragging the system down at start up.
You may also want to disconnect the EPS battery from the board (if it has one). Low battery voltage may also cause a loading problem.
If you don't have the Base System SW, you can find it in the manual at the link provided by dferguson. You'll want to have that on hand anyway, and the manual will help with minimum Config and troubleshooting. Not sure how much info that one provides. I've only used the proprietary service manual.
Hope this helps.
It wasn’t broken until I fixed it.
Report Post
Also, the full backup is a backup of the user defined settings, ie. presets, DICOM, reports, etc. It's not like a Ghost or Acronis backup of the full hard drive.
It wasn’t broken until I fixed it.
Report Post
Sorry, I'm not allowed to do that. I work for GE, and I can get in big trouble for sharing proprietary information. It should be in the manual that's available to you online. I'll take a look at mine and see what's in there.
It wasn’t broken until I fixed it.
Report Post
The part number for the Base System Software is 5135124-30 for all 2.x.x Application Software. I show none available to order through GE right now, but you might find it through a third party company.
It wasn’t broken until I fixed it.
Report Post
after verify error log i have these error in log évent
1) EchoVcr.GEdvrBoard with message DVR objet initiation failure
2) EchoVcr.GEdvrRecordUnit with message DVR Board not round
3) EchoVcr.GEdvrRecorderUnit with message failed to Stuart processus - net use 192.0.2.178/user root-
What means thèses errors?
Some one van help me
Report Post
my Logiq Book XP gives after start:
Operating Sysem not found. What is the way forward haven't try reloading severally INVALID DESTINATION is the result.
Anybody with solutions should plx reach out yemflab@hotmail.com
Tnks