Wed Sep 16 2020
philips epiq 5 error 503
Please help solve the problem
EventInfo InfoCategory="ApplicationEvent" LogCategory="Application" EventCategory="Error" EventID="600000.5MZ8.VTYA.3221225472" Description="FieldServiceText: SW:Most likely cause is that there was an I2C error.; StationName: ; Message: Failed to initialize ACB board TM1I2cDpu; Additional Details: fecErrorCode: 0, MoreInfo: timestamp: 0 usec;
Mon Sep 21 2020Reply from Alex156
Made TEST in logs - got two errors:
- Name="13 CB Neon1 To Xenon Receive" Result="Failed" Description="[ACB / Channel Bd]: Channel Board Neon1 to ACB Xenon Receive failure.
- Name="17 ACB PIC Status Test" Result="Failed" Description="[ACB] DPU Status Failure. EEPROM data CRC error occurred at last reset."
Who can tell you where on the channel Board is Neon1, closer to the ACB connector or further.
I think that the balls on Neon1 fell behind.
The problem appeared after removing the Board.
The second fault may be related to the first one.
Mon Sep 21 2020Reply from Ben C
Ben C
Agiliti
Conquest Imaging
Innovatus
Tue Sep 22 2020Reply from Alex156
 I really need firmware EEPROM ATH4502FCM from the ACB block share if anyone has I will be very grateful.
Tue Sep 22 2020Reply from Jim Carr
Jim Carr
Have you tried reloading software? The software installation will flash the firmware.
Best regards,
Jim Carr
AUS
USA office: +1 918-628-2831
Ego reinstalled luctus et error non abire
  8650752
  000
  000
  000
  1610
  000
  000
  000
Name="17 ACB PIC Status Test" Result="Failed" Description="[ACB] DPU Status Failure. EEPROM data CRC error occurred at last reset." Type="FunctionalTest"
Measurement ID="RSTResult" Name="ErrorCode" Result="Failed" LongValue 8650752
Measurement ID="0" Name="" Result="Failed"
Measurement ID="1000" Name="WriteI2C() result" Result="Passed"
Measurement ID="2300" Name="DPU status result" Result="Failed"